Author Topic: Atik 414 ex not cooling  (Read 8877 times)

emelan

  • Newbie
  • *
  • Posts: 1
Atik 414 ex not cooling
« on: July 29, 2020, 10:16:24 AM »
My Atik 424ex in not showing that it is cooling in Sequence Generator Pro. If I use Artemis then I can cool the camera down to -8 which is the delta -30 from my warm location 0f 24' c. If i cool the camera down to -8 or so the close Artemis and immediately open Sequence Generator Pro then the the camera cooler will show the -8 c but will not warm it up or cool it off any more. I am using the diver 6.4.0.0 and the atik camera chooser for ascom says Atik 414ex 118. I have also noticed a lot more hot pixels in my images lately. Is my camera on the way out of commission or is there a new driver i can use.

vince

  • Atik Staff
  • Sr. Member
  • *****
  • Posts: 283
Re: Atik 414 ex not cooling
« Reply #1 on: July 29, 2020, 11:39:44 AM »
Hello,

Are you running the latest Atik Core software from  https://www.atik-cameras.com/software-downloads/ ? 2020.06.23 is the latest version.

Best regards
Vince

emelan

  • Newbie
  • *
  • Posts: 1
Re: Atik 414 ex not cooling
« Reply #2 on: July 30, 2020, 01:05:51 AM »
Yes that did the trick. Thankyou

vince

  • Atik Staff
  • Sr. Member
  • *****
  • Posts: 283
Re: Atik 414 ex not cooling
« Reply #3 on: July 30, 2020, 10:16:43 AM »
Good to hear :)

Vince