Atik 460EX cooling issue
« on: September 19, 2020, 09:51:03 PM »
Having issues now with camera cooling. Last session I set to -10 and it never reached -10. After getting to around -7 or 8 at 100% power it stopped cooling. After disconnecting and reconnecting a couple of times I finally got it to reach -10 at about 96% power. This is a new problem I never had before. Camera is about a couple years old.
Thanks
Dave

vince

  • Atik Staff
  • Sr. Member
  • *****
  • Posts: 283
Re: Atik 460EX cooling issue
« Reply #1 on: September 21, 2020, 10:29:50 AM »
Hello Dave,

Sorry to hear you are having problems, what was your ambient temperature? Have you emailed Atik support with this issue?

Best regards
Vince

Hibou

  • Newbie
  • *
  • Posts: 4
Re: Atik 460EX cooling issue
« Reply #2 on: September 21, 2020, 05:19:29 PM »
The 460EX cools to -25C below ambient. So you will only get to -10C if ambient is 15C as Vince implies. Since this is a very low noise chip, you won't see the difference between -10C and -7C (or even 0C).